Your link is to a multitron not sensortron.

I saw the factory .pdf it says how to adjust but not where it is set from the factory.

The quarter rejection was very high, its in a vac so i brought the lid home and replaced the sensortron and it would not accept any coins, long story short i had the yellow ground hooked to the timed output of a gs-87, that was the problem.
 
Last edited:

Randy

Well-known member
Joined
Sep 5, 2007
Messages
5,859
Reaction score
2,211
Points
113
That’s weird I know I copied the Sensortron link and the Multitron link came up. Every Sensortron I've had to adjust the adjusting pot was in a different position from the factory. As the coin acceptor ages the settings will slightly change. There easy to adjust.
 

MEP001

Well-known member
Joined
Aug 30, 2007
Messages
16,665
Reaction score
3,946
Points
113
Location
Texas
There isn't a factory setting because the coils are all a little different. It has to be adjusted per unit. Basically you turn it down until it starts rejecting, then turn it back up just enough that it will take a range of years with no rejection.
 